Getting flow sensor frequency using FPGA and using DMA FIFO from target to host. Not getting correct number of samples in output TDMS file and unable to get correct number of pulses from flow sensor.
Hi GerdW,
There are 4 wires coming out of the sensor
brown to ni 9205 positive input
white is input of power supply positive
green to input power supply negative and to ni 9205 negative input.
I changed the power supply now to a 12 volt, 350 mA one and I am getting the output on oscilloscope as attached. This looks correct, is it?
Message 21 of 24
(902 Views)
Hi ash,
Where did you connect your scope?
When you are in Europe (or any other country with a 50Hz main supply) I would suspect you use a cheap power supply with no filtering. Or does your sensor accidentely output a 100Hz signal?
Message 22 of 24
(900 Views)
You were right, I checked the power supply, it outputs a signal with peaks of 100Hz instead of giving a dc voltage..
The way I connect the scope is, instead of the inputs going into the ni9205, I put in the same wires to the scope.
Message 23 of 24
(896 Views)
Hi GerdW... Yesterday I got a new flow sensor which is the same model. I now get proper pulses in the scope ..that means after everything.. the flow meter was defective🤦.. I will make the program now and hope for the best
Message 24 of 24
(887 Views)
