ni.com is currently undergoing scheduled maintenance.

Some services may be unavailable at this time. Please contact us for help or try again later.

LabVIEW

cancel
Showing results for 
Search instead for 
Did you mean: 

xy graph cutting data

Solved!
Go to solution

 I would check if this anomaly was really systematic, e.g. does this look always the same over a certain period of time e.g. one week?

 

 

2016-02-19_zoom.PNG

 

 

0 Kudos
Message 21 of 36
(3,052 Views)

I put a 1.5v (AA) battery on the 9215 and got this....Is there anyway I can calibrate this daq to make sure it's still within spec?

 

 

battery.JPG

0 Kudos
Message 22 of 36
(3,046 Views)

I'm on a laptop as well.

0 Kudos
Message 23 of 36
(3,044 Views)

Here's what the 1.5 looks like in MAX.

 

 

max.JPG

 

0 Kudos
Message 24 of 36
(3,031 Views)

That looks suspiciously like power line pickup.  How do you have everything connected, especially grounds?

 

Lynn

0 Kudos
Message 25 of 36
(3,000 Views)

What grounds do you mean? The 2 AI grounds are going into the AI- ports on the 9215. There's nothing special set up...I'm pretty sure it's the DAQ because my DOS scope reads the 2 channels fine.

0 Kudos
Message 26 of 36
(2,993 Views)

and for the AA battery I simply put the leads coming out of the DAQ in their respective polarities on the battery.

0 Kudos
Message 27 of 36
(2,991 Views)
Solution
Accepted by topic author glskinner84

Try connecting the source to COM as shown on Figure 5 (below) or Figure 6 of the NI 9215 Operating Intructions and Specifications document.

 

Lynn

 

 

 

Input connections.png

Message 28 of 36
(2,978 Views)

I'll post results of my attempt at your suggestion after lunch, John.

0 Kudos
Message 29 of 36
(2,920 Views)

It seemed to cleanup the signal but I'm still getting a weird offset.

 

1mohmdamping.JPGresistor.jpg

0 Kudos
Message 30 of 36
(2,909 Views)