 I would check if this anomaly was really systematic, e.g. does this look always the same over a certain period of time e.g. one week?

 

 

2016-02-19_zoom.PNG

 

 

0 Kudos
Message 21 of 36
(3,053 Views)

I put a 1.5v (AA) battery on the 9215 and got this....Is there anyway I can calibrate this daq to make sure it's still within spec?

 

 

battery.JPG

0 Kudos
Message 22 of 36
(3,047 Views)

I'm on a laptop as well.

0 Kudos
Message 23 of 36
(3,045 Views)

Here's what the 1.5 looks like in MAX.

 

 

max.JPG

 

0 Kudos
Message 24 of 36
(3,032 Views)

That looks suspiciously like power line pickup.  How do you have everything connected, especially grounds?

 

Lynn

0 Kudos
Message 25 of 36
(3,001 Views)

What grounds do you mean? The 2 AI grounds are going into the AI- ports on the 9215. There's nothing special set up...I'm pretty sure it's the DAQ because my DOS scope reads the 2 channels fine.

0 Kudos
Message 26 of 36
(2,994 Views)

and for the AA battery I simply put the leads coming out of the DAQ in their respective polarities on the battery.

0 Kudos
Message 27 of 36
(2,992 Views)
Solution
Accepted by topic author glskinner84

Try connecting the source to COM as shown on Figure 5 (below) or Figure 6 of the NI 9215 Operating Intructions and Specifications document.

 

Lynn

 

 

 

Input connections.png

Message 28 of 36
(2,979 Views)

I'll post results of my attempt at your suggestion after lunch, John.

0 Kudos
Message 29 of 36
(2,921 Views)

It seemed to cleanup the signal but I'm still getting a weird offset.

 

1mohmdamping.JPGresistor.jpg

0 Kudos
Message 30 of 36
(2,910 Views)