02-26-2019 03:55 AM
Hi,
we ordered a new one CDA 2990 clock. It can work either internal model or external model.
However, when the power on, only the power light indicator work, others indicator ( internal, external, pps) didn't work. When I connect a external CDA 2990 clock to the current on, the external, pps can work. The problem is this one can not work with internal model. But its a new one and its the first time to use. Please help me.
Bests,
Qu
Solved! Go to Solution.
02-28-2019 05:10 AM
Hi Qu,
Are you referring to the CDA-2990 or the CDA-2990 with GPSDO? Both of these products appear identical but only the version with GPSDO can generate clock and PPS signals internally by changing the switch to internal mode.
I can also recommend reaching out to Ettus Research's support page as this is the company that designed this hardware and therefore may have more insight as to what the issue is.
Kind regards,
Honn Son
Applications Engineer
03-06-2019 04:01 AM
Hi,
Thanks for your solution! Yes, I have verified with the centre yesterday, the clock we ordered is CDA-2990, not CDA -2990 with GPSDO.
Maybe we should consider take this one for exchange or order a new one with GPSDO.
Many thanks,
Qu