myDAQ analogue voltage measurement incorrect

Solved!
Go to solution

I am trying to read an analog signal from a sensor using the myDAQ. I could only manage to pick up a very faint signal. I then plug it to a signal generator, the 5V signal only reads as 88mV on the myDAQ. Has the DAQ gone bad? Thanks for any help.

 

IMG_20210928_180219.jpgIMG_20210928_180231.jpgIMG_20210928_180244.jpg

0 Kudos
Message 1 of 3
(1,178 Views)
Solution
Accepted by topic author samAAF

Hello,

 

As a test could you please connect the AI terminal to the +5V terminal and see the result (and nothing elce)?

Otherwise, I recommend you open an RMA request.

 

__________________________________________
The best way to thank, is to give KUDOS
Message 2 of 3
(1,144 Views)

It was the poor soldering to the BNC connector

Message 3 of 3
(1,073 Views)