‎07-10-2007 10:31 AM
‎07-11-2007
01:05 PM
- last edited on
‎07-20-2025
05:12 PM
by
Content Cleaner
Hello John,
The NI-6254 is capable of acquiring the data from your RVDTs. You would probably want to use all 16 analog differential inputs on the NI-6254. If all the RVDTs can be referenced to the same ground, then you could measure 32 analog inputs under the single ended mode. Differential mode is going to use two analog input channels to acquire data. They are paired up according to the following table.
A0 – A8 A16 – A24
A1 – A9 A17 – A25
A2 – A10 A18 – A26
A3 – A11 A19 – A27
A4 – A12 A20 – A28
A5 – A13 A21 – A29
A6 – A14 A22 – A30
A7 – A15 A23 – A31
The NI-6254 is going to have a range accuracy of 1.92mV with a Range of +10V to -10V. If you are getting a lot of noise with your signals connected in the single ended mode, you might want to try differential mode. Here is a great document that describes a few basic concepts on eliminating noise from your measurements. I would use the schematics in Table 1 of this document to connect a few biasing resistors to your circuit.
‎07-12-2007 09:39 AM
Hi Rob,
Thanks for you response. I decided to go a different way for the time being. The 6254 will be used in our final solution but it is just not accurate enough when I am measuring the two ac voltages calculating the rms and then the gain on the RVDT. I tried adding averaging and my numbers improved but we still were not where we needed to be. For this type of measurement we may have to move to an 18-bit card or even higher if they're available in the future. An 18 bit card would probably be good in the +/- 5 volt range. As I said we need the accurracy on our gain out to the ten thousandths place.
In our final solution, converting the two AC signals to a singal DC signal and calulating gain. The 6254 will be more than ample enough for the job.
Thanks for your help!
Regards-
‎07-12-2007 11:24 AM
‎07-13-2007 09:58 AM