error code -200081

I am trying to calibrate a NI 9211 unit but I keep getting the following error code.

 

Error or termination occurred. This calibration may not be valid. Error code: -200081 Error message: Error -200081 occurred at DAQmx Start Task.vi at step Verify Analog Input:1 Number of Channels: 2 Sample Rate: 7.140001 Maximum Sample Rate: 6.000000

I am using CalExec version 3.3.2. Is there a fix to this problem?

 

 

Ryan A. Saunders

Idaho National Laboratory

Standards and Calibration Lab

phone: (208)526-2202

email: Ryan.saunders@INL.gov

0 Kudos
Message 1 of 5
(2,621 Views)

Hi Saipanman,

 

Which DAQ Chassis are you using??

Are you are using a single Module Carrier such as a NI USB-9162 or are you using a cDAQ chassis such as the NI cDAQ-9172?

 

  

Nathan
Applications Engineer
National Instruments UK & Ireland
0 Kudos
Message 2 of 5
(2,605 Views)

The user has it in a NI USB-9162

0 Kudos
Message 3 of 5
(2,599 Views)

Hi Saipanman,

 

Calibration is only possible will a Chassis such as the 9172 as this is not available for the single module carrier such as the 9162.

If you are able to, try using a Chassis to perform the calibration instead.

Nathan
Applications Engineer
National Instruments UK & Ireland
Message 4 of 5
(2,585 Views)

Wilson54910

 

Great info Smiley Happy

 

 

 

thanks for your help.

 

Totally Kudos

0 Kudos
Message 5 of 5
(2,566 Views)