NI 9224 open circuit voltage reading is about 1.47 V

Solved!
Go to solution

Hi everyone, can't seem to find anything about this around the web so I figured I'd make a forum post.

 

I have a cRIO 9045 and a couple NI 9224 AI voltage modules. They work as you'd expect when hooked up to a voltage source and when you short a channel, but things get weird when it's an open circuit. For some reason the open circuit voltage reading is ~1.47 volts and I've never been able to make any sense of it. We have another cRIO with a few NI 9224 modules, and my coworker told me that he's seen the same thing on that one.

 

I don't think it has anything to do with the DAQmx code, but here's how the channels are configured just in case:

rseabeck_0-1673542952780.png

Additionally, here is 18 different AI voltage inputs on our other cRIO reading about 1.47 volts at open circuit to demonstrate that it's repeatable and consistent.

rseabeck_0-1673557755578.png

 

Anyway, thanks for any suggestions. If we have to live with this then it's not the end of the world, but it would improve our error handling capabilities a lot if we can figure this out.

0 Kudos
Message 1 of 4
(2,003 Views)
Solution
Accepted by topic author rseabeck

NI does not specify what voltage a device will read when there is no signal connected. See Incorrect Readings on Unconnected or Open Channels of DAQ Device

-------------------------------------------------------
Applications Engineer | TME Systems
https://tmesystems.net/
-------------------------------------------------------
https://github.com/ZhiYang-Ong
0 Kudos
Message 2 of 4
(1,985 Views)

@ZYOng wrote:

NI does not specify what voltage a device will read when there is no signal connected. See Incorrect Readings on Unconnected or Open Channels of DAQ Device


Huh, so I guess the solution really is to just live with it. Thanks for the link!

0 Kudos
Message 3 of 4
(1,959 Views)

This is a common issue with all multiplexed signal systems so I'm surprised your searches came up empty.  It's called "ghosting" and there are ways to deal with it.

 

How Do I Eliminate Ghosting from My Measurements? - NI

 

Most important though, is the possibility of actually incorrect readings due to cross-talk when jumping quickly between a large voltage and a small voltage on adjacent channels.  If you scan too quickly, the multiplexor and instrumentation amp will still hold a charge from the large signal and you will read it as an error in the small signal channel.  You can adjust the inter-channel timing to fix it or scan "dummy" channels that are grounded between each real channel.

 

It's all in here:  Getting Started with NI-DAQmx - NI

LabVIEW Pro Dev & Measurement Studio Pro (VS Pro) 2019
Message 4 of 4