ni.com is currently undergoing scheduled maintenance.

Some services may be unavailable at this time. Please contact us for help or try again later.

Signal Conditioning

cancel
Showing results for 
Search instead for 
Did you mean: 

SCXI 1520 Simultaneously sampling

Hi there,

for my application i need to read simultaneously all the channels from 2 SCXI 1520 modules. (2 x 8 = 16 channels)

The control module will be SXCI 1600 and the  software  -  Signal Express. Is this possible ?

0 Kudos
Message 1 of 3
(6,682 Views)

........probably this is not possible. From what i've found on NI site, 2 SCXI 1520 modules should run in Parallel mode and each of them requires a separate DAQ board .

DAQ boards can be synchronised through a single RTSI cable. Any suggestions about the synchronization of the DAQ boards in Signal Express ?

0 Kudos
Message 2 of 3
(6,677 Views)

Hello metrisys,

 

I believe that it is possible to simultaneously sample from analog input channels on two SCXI-1520 modules that are controlled with the SCXI-1600.  PFI 0 of the SCXI-1600 is used to transmit the signal that triggers an acquisition for the SCXI-1520s.  You would need to configure the acquisition to recognize the signal from PFI 0 of the SCXI-1600 as the start trigger, and add all of the channels on the two SCXI-1520s to the scan list of your analog input task. For your convenience I've posted a link to a knowledge base article titled 'Synchronizing Multiple SCXI Simultaneous Sample-and-Hold (SSH) Modules.'  

 

 http://digital.ni.com/public.nsf/allkb/A01F8168AFCFE9DC86256B3300758CD0?OpenDocument

 

Best wishes,

Wallace F.

National Instruments
Applications Engineer
0 Kudos
Message 3 of 3
(6,657 Views)