03-01-2021 12:45 AM
Hi everyone,
If I need a strain measurement data point from the specimen for every 1 microsecond then I need a measuring device with a frequency bandwidth of 1000KHz and above?.
But the catalog for strain instruments (C Series Strain/Bridge Input Module) has a maximum of 50KS/s/ch. This can give a measuring point every 20 microseconds, which I think is not suitable for dynamic material characterization and impact experiments.
Can someone tell me what devices are used for SHPB and impact data measurements form NI.
03-18-2021 01:57 AM
Hello Bhaskarhnk,
The PXIe-4331 has a maximum sample rate of 102Ks/s, which can allow you to strain measurement data point from the specimen for every 10 microseconds
03-19-2021 05:12 PM - edited 03-19-2021 05:17 PM
You need 1MSPS and assuming a <0.5% uncertainty at ~100 mV ...
We use a PXI 5922 for high shock calibration, but we use an additional signal conditioner only up to 300kHz,. but a good 1MHz bridge amplifier ??? would be challenge to calibrate it 😉
So a 1MSPS (an no multiplexing!) DAQ with external amplifier looks reasonable ... unless a NI-supporter comes up with a 1MSPS DAQ and less than 2mV uncertainty 😉
And I want NI product finder with a general spec search including spec'ed uncertainty, min max ranges, but maybe I missed it 😉
06-10-2021 04:27 AM
Hi Bhaskar,
I am using PXIe-6124 card (4MS/sec) for SHPB.