PXIe-6570 Force/Measure Current Accuracy Computation

Solved!
Go to solution

Hi,

 

I am having problems on understanding the computation for PXIe-6570 Force/Measure Current Accuracy.

What comes next after finding the zone (step 6)?

Can anyone give me an example of the computation?

 

Thanks!

Download All
0 Kudos
Message 1 of 3
(965 Views)
Solution
Accepted by trixx25

Let me show you an example for calculating accuracy,

 

Steps

  1. FI = 1.5mA
  2. Using 2mA as the best range
  3. % Force current range = (1.5m / 2m)*100 = +75%
  4. Assume my impedance is 1k Ohm, Voltage Required = 1k * 1.5m = 1.5V
  5. Plotting the 1.5V, +75% in the Figure3, the point lies in the grey area which is designated as zone 1
  6. Looking at Table1, it states that +-1% for Zone1 irrespective of range
  7. Though not explicitly stated, you calculate the accuracy as you typically do for any measurement or value in engineering, apply +-1% to +-2mA range, now it becomes +-40uA because +-2mA constitutes 4mA total range.

Hope step 7 clarifies your question about how to apply +-1% to get the accuracy number.

 

Alternatively, if my impedance was 100 ohm and force current was -1.5mA, the voltage required becomes 150mV. Now, the 150mV, -75% lies outside zone1 in the graph and hence there is no warranted accuracy specification (as mentioned in step5)

Santhosh
Soliton Technologies

New to the forum? Please read community guidelines and how to ask smart questions

Only two ways to appreciate someone who spent their free time to reply/answer your question - give them Kudos or mark their reply as the answer/solution.

Finding it hard to source NI hardware? Try NI Trading Post
Message 2 of 3
(931 Views)

It makes so much more sense now.

Thank you!