memco,
First of all, how are you connecting your analog signal to the PXI-6120? If your signal source is coming from the PXI-5411, then this is a grounded signal source, which requires connection to the AI<X>+ and AI<X>- terminals of the 6120. This is shown in the 6120 User Manual linked below:
NI 6115/6120 User Manual
For example, if you were measuring on channel 1 of the 6120, your connections would be to pins 33 and 66, which correspond to AI1+ and AI1-. You should first check to see if you get a valid reading in Measurement and Automation Explorer (MAX) before trying to develop an application in LabVIEW or other programming language.
03-08-2006 09:54 AM
I have the outputs of the 5411 connected (5V @ 2000 Hz)
Center BNC -> Pin 30
Shield BNC -> Pin 63 (or Pin 29 same result)
Reading Dev9/ai3 0.325 Volts.
And a scope connected to Pin 30 to observe to signal which reads correctly!
I have tried other AI channels with the same result.
If I apply a correction factor of 11.318 = (8*sqrt(2)) to the amplitude I get good
agreement over a range of 0-7 Volts.
Also I measure the correct frequency.
Ernest
03-09-2006
06:36 PM
- last edited on
08-18-2025
04:04 PM
by
Content Cleaner
03-13-2006 08:49 AM
03-16-2006 09:03 AM
03-17-2006 09:14 AM
memco,
Well those measured voltages certainly do not look good. One thing you can try to do as well is try a loopback type test and run one of your analog output channels directly to an input. In this case, we know that the ground reference will be the same. Also, going back to your original issue measuring the output of the 5411. Did you try changing the load impedance of the 5411? If you are using the NI-FGEN Soft Front Panel, this can be changed in Edit >> Device Configuration. Also, how old is the 6120? Do you know when it was last calibrated?
Regards,
LA