I have the outputs of the 5411 connected (5V @ 2000 Hz)
Center BNC -> Pin 30
Shield BNC -> Pin 63 (or Pin 29 same result)
Reading Dev9/ai3 0.325 Volts.
And a scope connected to Pin 30 to observe to signal which reads correctly!
I have tried other AI channels with the same result.
If I apply a correction factor of 11.318 = (8*sqrt(2)) to the amplitude I get good
agreement over a range of 0-7 Volts.
Also I measure the correct frequency.
Ernest
03-09-2006 06:36 PM
03-13-2006 08:49 AM
03-16-2006 09:03 AM
03-17-2006 09:14 AM
memco,
Well those measured voltages certainly do not look good. One thing you can try to do as well is try a loopback type test and run one of your analog output channels directly to an input. In this case, we know that the ground reference will be the same. Also, going back to your original issue measuring the output of the 5411. Did you try changing the load impedance of the 5411? If you are using the NI-FGEN Soft Front Panel, this can be changed in Edit >> Device Configuration. Also, how old is the 6120? Do you know when it was last calibrated?
Regards,
LA