USB 6009 not responding in Lab VIEW, Analog Input

Solved!
Go to solution

Hi, 

 

So I am trying to create a task using the Analog Input found in the USB-6009, however, says that there are "no supported devices"

I am really confused because I can see the USB-6009 in my NI MAX, and I can create tasks for digital input. 

 

I read that I need to "Make sure that the task type is supported by the module. For instance, a digital module cannot support an analog task." at https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z000000P7OaSAK&l=en-US

but I can't find any further documentation on what specific steps to take.

I am also unsure whether this is even my problem.

 

I am new to LabVIEW, so any advice you could give I would really appreciate it.

0 Kudos
Message 1 of 4
(782 Views)

Can you share the code you used to reproduce the error? 

0 Kudos
Message 2 of 4
(770 Views)

I have attached an image of what I am talking about. 

So the issue occurs when I try to use the DAQ-Assist function in the block window.

 

Thank you for your time.

0 Kudos
Message 3 of 4
(753 Views)
Solution
Accepted by KhelfH

Strain measurement requires excitation voltage. USB-6009 does not support it.

It is recommended to use the specialized strain input module.

Measuring Strain with Strain Gages

C Series Strain/Bridge Input Module