11-07-2006 01:50 AM
11-08-2006 01:20 PM
12-06-2007 05:58 PM
I had similar problem, and using the suggestion, I double checked all the connectiong and I am still getting 289°C value using a K type TC. I know the lab is not that hot!
It had worked at one point, but as I was teting other part of the LabVIEW code, it all of a sudden stopped working and started to give me very strange numbers in the 289 range
any other thoughts?
thanks!
12-07-2007 10:30 AM
Hi NewUser2007,
You said: “It had worked at one point, but as I was teting other part of the LabVIEW code, it all of a sudden stopped working and started to give me very strange numbers in the 289 range.”
Do you mean:
1. Your code was executing fine and then you get 280°C forever afterward even after stopping and running your code again? or
2. Running your code gives good results until a certain point every time it’s run?
In other words, did a connection come loose and now you are not receiving the signal properly because of the hardware, or did something change in your software that altered the expected functionality?
For hardware, I would recommend reading through the SCC-68 User Guide. It is a pretty thorough resource for setting up the hardware correctly for your particular hardware.
If you suspect your code to be at fault, try using a shipping example (From LabVIEW, select Help » Find Examples and search for Thermocouple examples).
If you are still having difficulty, please outline your entire setup so that all variables of your system are considered.
Additional resources:
Taking Thermocouple Temperature Measurements tutorial at ni.com/zone