 the_dani
			the_dani
			07-14-2017 07:37 AM - edited 07-14-2017 07:38 AM
I'm trying to measure the output current of a sensor. It is between 4mA and 20mA.
I'm using the NI cDAQ-9171 with the NI 9218 Unit and an additional shunt resistor, NI 9983F (for +- 20mA current measurement):
In the "Start Task" vi does an error occur:
Error -200077 at DAQmx Start Task.vi: 7220001
Property: AI.Max
Requested Value: 20,0e-3
Value must be bigger than: -6,644518e-3
Value must be smaller than: 6,644518e-3
Device: cDAQ1Mod1
Task-Name: _unnamedTask<44>
I tried several examples made by NI or other users but nothing worked!
My current measurement vi.
Thank you in advance!
 Duffy2007
			Duffy2007
			07-17-2017 02:55 AM - edited 07-17-2017 02:58 AM
Hello the_dani,
which driver you are using with DAQmx 16.1 and newer this should not happen, with an older driver the worlkaround is to configure the adapter to NI 9983D.
07-18-2017 01:05 AM
Thank you, I updated DAQmx and it works!