09-02-2024 08:27 AM
Hi,
I have a new USB-6000 device.
I am trying to connect sensors but from the output, the voltages do not change when I attach an accelerometer or thermocouple or any sensor at all.
When I shake the accelerometer, the voltage doesn't reflect the change.
it is almost as if the device isn't picking up the sensors at all.
The only reading I am getting in analog input is the base voltage reading.
I am pretty sure the sensors are working as they worked on an old device.
What could be the issue.
Thanks
09-02-2024 11:15 AM
You need to explain in details (with photo, connection diagram, VIs etc.,)
09-02-2024 11:15 AM
Are you connecting your sensors differential way or single ended , check which one is correct for your connection
thanks
09-03-2024 02:19 AM
I am using single ended as the DAQ device only accepts single ended.
The software i use is daq express
09-03-2024 02:45 AM
The sensors needs an external power source as the DAQ device doesn't power up the sensors.
I have attached the schematic diagram of the accelerometer.
connected the output axis to channels 0-2
ground and 10v power supply from an external source.
attached the daq express software photo showing the voltage outputs (these are the same outputs i get when no sensor is attached as well)
Thanks
09-03-2024 06:21 AM
Okay I have been able to figure out the accelerometer but how do I make the type k thermocouple work and actually measure voltage changes when i place the thermocouple in hot water.?
the attached picture is from the daq device datasheet.
I have attached the full datasheet for the usb-6000 daq device.
09-03-2024 07:57 AM
@ppaulade wrote:
Okay I have been able to figure out the accelerometer but how do I make the type k thermocouple work and actually measure voltage changes when i place the thermocouple in hot water.?
the attached picture is from the daq device datasheet.
I have attached the full datasheet for the usb-6000 daq device.
USB-6000 specifications is insufficient to measure a Type-K thermocouple reliably.
With 26-135mV error over temperature and AI range, it covers the entire working range of Type-K Thermocouple.
Type-K TC generates -6.458mV(-270 degC) to 54.886mV (+1373 degC) as per this reference.