Data Accqusition Error

I have a 6062E card and I am accquiring the data at 15Khz and I want to accquire data for 30sec which tranlates to 450,000 samples. I am gathering data from all the 8 channels at this rate.
 
When I try and write the data to the file It writes only 150,000 samples/ch.
 
I do get the error number :200284.
 
Is there a limitation on the card with respect to buffer size for a single accqusition?
 
Thanks
 
Prasad
 
0 Kudos
Message 1 of 3
(3,138 Views)
The read would be expected to take about 30 seconds to complete, but instead an error is returned after about 10 seconds.  Many of the DAQmx examples have a read timeout value of 10 wired to DAQmx Read.  Checking this value and setting it to something higher than 30 or -1 for an infinite timeout will likely solve your problem.

Your board has a limited onboard buffer, but the size of the buffer on the PC performing the acquisition is only limited by the system's capabilities.

I hope this helps.

Jason
Message 2 of 3
(3,131 Views)

Jason

Thanks for the help. That seems to have fixed the problem. My understanding was that -1 would be the default so I had left that unwired.

 

Thanks again

 

Prasad