ni.com is currently undergoing scheduled maintenance.

Some services may be unavailable at this time. Please contact us for help or try again later.

Multifunction DAQ

cancel
Showing results for 
Search instead for 
Did you mean: 

Comparing Measurement Computing USB 2416 vs NI9215

We're considering using cDaq for an application that  will read a voltage signal of approx 0.26 to 0.36 V. 

 

I did a test with a measurement computing USB 2416 -24 bits resolution and compared with NI 9214 4 16bit simultaneous AD.

 

For both converters I used the following parameters:
S/R - 500 S/s

Range -+10V (just because NI9215 only accepts that range so I used this in both for fear comparison)

Nbr of Samples - 2500 samples - 5 seconds of data. 



The results are shown in the screen shot bellow. 

I subtracted the mean from all the array and added all the "noise" together. For the sensor at 0.22V the sum was 

USB 2416 - 0.04

NI9215 - 13.673

 

The graphic shows the histogram distribution of the errors in percentage. 

 

I understand that the NI9215 may not the ideal for my application, my goal is to use a NI 9205 (32 Inputs 16 Bits with -+1 V) but my concern is about the noise level on this board. 

Can anyone from NI to give me some guidance if the NI9205 will have better performance ? Am I missing something ? 

I tried different sources, one precision power source at 0.7V and 5V signal from USB2416. at 5V they are equivalent but bellow 1V it looks very noisy. 

I tried to borrow NI cards for test but it's not available. 

I appreciate all inputs/comments. 

 