Multifunction DAQ

cancelar
Mostrando resultados para 
Pesquisar então 
Você quer dizer: 

Calibration Executive

Why do the specifications differ between the Cal Exec calibration reports and the specs on NI's website? How where the calibration executive specifications determined?

0 Kudos
Mensagem 1 de 5
3.614Exibições

Hello,

 

I can not look into this without knowing the following:

 

1.) What product are you using?

2.) Which specification isn't matching up and the calibration procedure that isn't working properly?

3.) What information are you comparing? Could you provide the documents that you are comparing?


Jim St
National Instruments
RF Product Support Engineer
0 Kudos
Mensagem 2 de 5
3.608Exibições

Do you know if NI has warranted specifictions for the Digitizer cards? For example, the PXI-5154 has 'typical' for the RMS noise, Rise/Fall Time, and Crosstalk specifications on the spec sheet. Are there warranted specifications anywhere?

 

Thanks

 

0 Kudos
Mensagem 3 de 5
3.480Exibições

There aren't any warranted specifications available for our typical specs on digitizers.  It must be listed as warranted in the manual.

 

Jon S

Applications Engineer
National Instruments
0 Kudos
Mensagem 4 de 5
3.411Exibições

Jim,

 

Thanks for the response, however after looking at the specifications it seems that the calibraiton executive specifications are always tighter than the NI spec sheet, so even though they don't match we are okay as long as they are tighter specs.  Thanks

0 Kudos
Mensagem 5 de 5
3.407Exibições