 LudovicR
			LudovicR
			09-24-2025 09:16 AM
Dear All,
We have an analogue-controlled XY scanner equipped with a PCI 6711 and a BNC-2110. During a raster scan, artefacts appear in the Y scan (the ramp is broken) at certain X positions (around 0V). I put the signals on the oscilloscope. Changing the cables has no effect; the problem appears to originate from AO1.
Could you advise me on how to identify the source of the problem?
Ludovic R.
 santo_13
			santo_13
			09-24-2025 10:17 AM
Source of the problem is clearly your code that generates the ramp signal and definitely doesn't look like DAQ (if you assumed so)
 altenbach
			altenbach
			09-24-2025 11:23 AM - edited 09-24-2025 11:26 AM
09-25-2025 04:26 AM
Thank you very much for your advice.
It is Windows 10 and Labview 2011.
I have attached the vi. It is an old code (< 15 years old) and has not caused any problems until now.
As a result, I had not made any changes to it. But it is clearly very basic, which accentuates the problem.
The connection to the oscilloscope is direct: PCI6711-> BNC 2010-> BNC cable -> oscilloscope.
What is strange is that the artifacts in AO1 (4 ramps over 256) are correlated to the voltage in AO0, regardless of the scanner configuration.
At this time/point, this is controlled by the software.
Best regards
09-25-2025 04:27 AM
Vi and sub vi
 Henrik_Volkers
			Henrik_Volkers
			09-25-2025 04:46 AM
You have the code
it was running flawless in the past
What has changed? Only the output ? Something else?
(the PCI card rely on the PCI bus .. DMA etc .. other software/tasks running? )
I would swap ouput channels in software (and change the cabling accordingly) to see if the problem stay with the hardware channel or the software channel.
 altenbach
			altenbach
			09-25-2025 12:44 PM
I agree that the code is probably OK, but I had a quick glance at some of your VIs and if you ever get a chance, some could be dramatically simplified.
For example the raster could be done more easily as follows. (I am not 100% sure if LabVIEW 11 had the concatenating tunnel, but here's how you could write it today. much less clutter and fewer places for bugs to hide.
I am pretty sure that your ellipse (which is a circle because you ignore the y) is buggy, because it only converges to a circle for very high point counts. As you can see, the actual radius fluctuates because one of the shift registers starts with a non-empty array, so the first point is an "almost" duplicate of the second point. (Note that if you don't want to close the circle, you can change N and set "exclude end=true" for the ramp.)
A circle is most easily done using a complex datatype. The tiny code on top does the same (and better!) as your entire subVI.
(Most of your other code could be corrected similarly.)
09-26-2025 05:34 AM
Thank you very much, Altenbach, for looking at the code and for your advice. I will make the changes and check to update the rest of code.
I will see what impact they have.
Regarding past events on the machine, I had a broken hard drive this spring.
Replacing it with the backup worked fine, but there may have been a bug during the update process, although that seems unlikely.
Best regards