Artefact in Analog signal
Dear All,
We have an analogue-controlled XY scanner equipped with a PCI 6711 and a BNC-2110. During a raster scan, artefacts appear in the Y scan (the ramp is broken) at certain X positions (around 0V). I put the signals on the oscilloscope. Changing the cables has no effect; the problem appears to originate from AO1.
Could you advise me on how to identify the source of the problem?
Ludovic R.
Message 1 of 8
(439 Views)
Source of the problem is clearly your code that generates the ramp signal and definitely doesn't look like DAQ (if you assumed so)
Santhosh
Soliton Technologies
Message 2 of 8
(419 Views)
  • We would need to see the code that generates these ramps.
  • You seem to have 20us/point. Is this hardware or software timed?
  • I assume this is running in windows, not a RT OS.
  • Why is the signal generally so noisy? Are things properly terminated?
Message 3 of 8
(404 Views)
Thank you very much for your advice.
 It is Windows 10 and Labview 2011.
I have attached the vi. It is an old code (< 15 years old) and has not caused any problems until now.
As a result, I had not made any changes to it. But it is clearly very basic, which accentuates the problem.
 The connection to the oscilloscope is direct: PCI6711-> BNC 2010-> BNC cable -> oscilloscope.
 What is strange is that the artifacts in AO1 (4 ramps over 256)  are correlated to the voltage in AO0, regardless of the scanner configuration.
 At this time/point, this is controlled by the software.
Best regards
Message 4 of 8
(362 Views)
Vi and sub vi
Message 5 of 8
(361 Views)
You have the code
it was running flawless in the past
What has changed? Only the output ? Something else?
(the PCI card rely on the PCI bus .. DMA etc .. other software/tasks running? )
I would swap ouput channels in software (and change the cabling accordingly) to see if the problem stay with the hardware channel or the software channel.
Greetings from Germany
Henrik
LV since v3.1
Message 6 of 8
(354 Views)
I agree that the code is probably OK, but I had a quick glance at some of your VIs and if you ever get a chance, some could be dramatically simplified.
For example the raster could be done more easily as follows. (I am not 100% sure if LabVIEW 11 had the concatenating tunnel, but here's how you could write it today. much less clutter and fewer places for bugs to hide.
I am pretty sure that your ellipse (which is a circle because you ignore the y) is buggy, because it only converges to a circle for very high point counts. As you can see, the actual radius fluctuates because one of the shift registers starts with a non-empty array, so the first point is an "almost" duplicate of the second point. (Note that if you don't want to close the circle, you can change N and set "exclude end=true" for the ramp.)
A circle is most easily done using a complex datatype. The tiny code on top does the same (and better!) as your entire subVI.
(Most of your other code could be corrected similarly.)
Message 7 of 8
(303 Views)
Thank you very much, Altenbach, for looking at the code and for your advice. I will make the changes and check to update the rest of code.
I will see what impact they have.
Regarding past events on the machine, I had a broken hard drive this spring.
Replacing it with the backup worked fine, but there may have been a bug during the update process, although that seems unlikely.
Best regards
Message 8 of 8
(287 Views)
