This setup is using a USB-6216 with vb.net using the most recent nidaqmx 9.40.
I'd like to synchronize a PWM waveform (say 30 Hz, 10% duty cycle) with the analog input on one channel (61440 ksamp/s on AI0). Steps taken are :
--
aiTask.AIChannels.CreateVoltageChannel("Dev1/ai0", "", AITerminalConfiguration.Rse, -10, 10, AIVoltageUnits.Volts)
ctrTask.COChannels.CreatePulseChannelFrequency("Dev1/ctr0", "", COPulseFrequencyUnits.Hertz, COPulseIdleState.Low, 0, 30, 0.1)
aiTask.Timing.ConfigureSampleClock("", 61440, SampleClockActiveEdge.Rising, SampleQuantityMode.ContinuousSamples,61440)
ctrTask.Timing.ConfigureImplicit(SampleQuantityMode.ContinuousSamples, 1024)
--
However, I see a large drift on the waveform acquired - the square waveform is drifting later (on the right on a time domain graph) by approximately 100 samples per second. Playing with the sampling rate will change the drift to higher or lower amount, but should I expect the waveform to be timed with the analog input? Is there a way to synchronize the timebases for both of the tasks?
I've tried to look into the examples in the DotNet catalog, however, it seems the Multi-Function example is not supported by the USB-6216.
Thanks, JC
08-12-2011 11:42 AM
JC,
In this situation, what I would do would be to generate your signals from the same clock would be to generate a 30 Hz continous pulse train on one counter and the generate a 61440 Hz continuous pulse train on the other counter with a 50% duty cycle. I would then use the 61440 Hz pulse train as your sample clock for your AI. This would ensure that they are on the same clock. How ever you might see a little drift because 61440 Hz doesn't divide into 80 Mhz evenly, so the clock wont be exactly at 61440 while 30 does divide evenly. Take a look at this discussion forum. The fourth post down explains how to do this.
I hope that helps.
Brian P.