Hello sir:

Regarding "Simple Calibration.vi", i had a question:

捕获.PNG

why they use "0.15" for "IMAQ set simple calibration2 vi"? it mean 1pixel=0.15Millimeter? but actually i got 1pixel=0.083Millimeter from picture's resolution. i don't understand these relationship. is there any help? thanks.

 

 

0 项奖励
11 条消息(共 34 条)
3,075 次查看

 

 

0 项奖励
12 条消息(共 34 条)
3,077 次查看

Time_Walker 已写:

Hello sir:

Regarding "Simple Calibration.vi", i had a question:

捕获.PNG

why they use "0.15" for "IMAQ set simple calibration2 vi"? it mean 1pixel=0.15Millimeter? but actually i got 1pixel=0.083Millimeter from picture's resolution. i don't understand these relationship. is there any help? thanks.

 

 


any help?

0 项奖励
13 条消息(共 34 条)
3,063 次查看
Yes, based on your image setup and calibration grid image it can vary.
-Yes it means one pixel is equivalent to 0.15mm in X direction and 0.15mm in Y direction
Thanks
uday
0 项奖励
14 条消息(共 34 条)
3,062 次查看

udka 已写:
Yes, based on your image setup and calibration grid image it can vary.
-Yes it means one pixel is equivalent to 0.15mm in X direction and 0.15mm in Y direction

but actually i got the result from picture's resolution :

the resolution for picture is 300dpi. it mean 300pixels/inch, 1inch=25.4Millimeter.

300pixels/25.4Millimeter.

1pixel=0.083Millimeter.

why it is so huge different with "0.15mm" specified in picture? "0.15mm" is an arbitrary setting for demostration? when we need to set "X step", "Y Step" parameter in application, how to caculate it and set it?

0 项奖励
15 条消息(共 34 条)
3,057 次查看
you have to use sample target in your image that you know the dimensions of this sample
0 项奖励
16 条消息(共 34 条)
3,037 次查看

one more questiion:

X Step and Y Step:

"Scaling factors are the real-world distances between the dots in the calibration grid in the x- and y- directions and the units in which the distances are measured. Use the X Step and Y Step elements of the Grid Descriptor control to specify the scaling factors." this is the definition in "https://zone.ni.com/reference/en-XX/help/370281M-01/nivisionlvbasics/perspective_and_nonlinear_distortion_calibration/ "

 

the following is the definition in the labview help for "IMAQ Calibration Target to Points - Circular Dots VI", "X Step corresponds to the width of a pixel in real-world units." it make me completely confused.

捕获.PNG

 

Any help ? which one is right? Any official explaination? thank you very much ~~~~


Hatef.fouladi 已写:
you have to use sample target in your image that you know the dimensions of this sample

 

0 项奖励
17 条消息(共 34 条)
3,019 次查看
witch part confused you?
0 项奖励
18 条消息(共 34 条)
2,992 次查看

Hatef.fouladi 已写:
witch part confused you?

X Step and Y Step:

"Scaling factors are the real-world distances between the dots in the calibration grid in the x- and y- directions and the units in which the distances are measured. Use the X Step and Y Step elements of the Grid Descriptor control to specify the scaling factors." this is the definition in "https://zone.ni.com/reference/en-XX/help/370281M-01/nivisionlvbasics/perspective_and_nonlinear_disto... "

 

the following is the definition in the labview help for "IMAQ Calibration Target to Points - Circular Dots VI", "X Step corresponds to the width of a pixel in real-world units." 

捕获.PNG

 

The definition for X-step and Y-Step are the same? or they are different? i marked it with red color~

19 条消息(共 34 条)
2,989 次查看
解答
已被主题作者 Time_Walker 接受

sppose you have two point  with distance of 1 cm 
you  find out that this two point distance is 3 pixcel in image 
so if you check bellow image you have this 
x step 1
y stem 0,98
unit :cm