Hello all,
I got a problem with my stand alone application in the target computer.
My program has a sub vi named Igenerator, which generates current with the connected NI 9265 with all 4 channels.
When building the installer of the application, I already included NI-DAQmx application development support, and with this
installer my application could be opened in the target computer correctly.
however when running the application in my target computer (with NI 9265 connected), from the result of the program I can
see that there's no current output from NI 9265, so it is an issue of transfer the DAQ settings from my original computer where
the project was generated to my target computer, isn't it?
To solve the problem I found this link: http://www.ni.com/white-paper/5406/en, seems that the trick is in the last step of building
the installer I should include a hardware configuration. But how? this link fails to clarify this point to me, and I tried to do 'configure',
and export all my current generation task from MAX to the file 'configdata.exe', and I clicked 'Build', it seemed fine, but in my target
computer after I setup the installer, my application again has no current output from NI 9265.
Someone got any idea about how could I achive that? And any mistake of my procedures I've done? thanks
Chao
Solved! Go to Solution.
10-09-2013 10:31 AM - edited 10-09-2013 10:36 AM
[Edit] I just learned something. Prior data was incorrect.
Does the configuration appear correct in MAX after the installer runs on the target machine?
10-09-2013 10:48 AM
Well I'm not sure if I got your question. I set up the installer successfully in the target computer and the stand alone
application runs well. But i don't know how to check the configuration in my installer with MAX.
Yes I did export the current tasks from the MAX in my original computer to the installer, but how could
I check that in MAX in my target computer?
10-09-2013 11:00 AM - edited 10-09-2013 11:05 AM
Just a thought...
did you check that the device name is the same in MAX in both the target and the developpement computer?
regards
antoine
edit: devive name is the one between "" character in max, like "cDAQ1"
10-09-2013 11:00 AM
Open MAX on the target PC after the installer runs.
Do Tasks, System devices and Scales match the setting from the development PC?
10-09-2013 01:27 PM
Ok I got it. After I reset the configuration in MAX and re-installed the installer, I checked MAX, and found that the
NI-DAQmx tasks are the same as they are in my original computer, but under the option 'Device and interfaces' I
see 2 NI 9265! 'Dev 1' and 'Dev 2', one has green icon and one has yellow icon on the left. And it shows that my
PC could communicate successfully with both!
Anyway this time I failed again, after checking MAX only Device part is odd. So any idea?
10-09-2013 01:28 PM
Hi Antoine, my reply is above as it's also for Jeff.
Thanks.
10-09-2013 01:30 PM
Hello jcraffael,
The yellow icon indicates that the device is simulated rather than physically present. You'll need to delete this device and rename the physical device to match the configured task.
Regards,
10-09-2013 02:11 PM
One small problem with Simulated devices:
You Cannot assign a new real device to the alias of a simulated device.
See Here- I probably should have reported this as a bug a long time ago
The workaround is obvious: Clear simulated devices from targets before installing applications that need that alias or running a system configuration API utility to assign a application specific alias
10-10-2013 03:23 AM
Hi Tom! Thanks for the explanation. I deleted the device with yellow icon 'Dev 1', and renamed the device with green icon 'Dev 1' from 'Dev 2', and the system asked me to assign the tasks related to the old 'Dev 1' to the new 'Dev 1' and I did it.
However, it still didn't work. I think probably I made mistake in implementing the procedure above....