05-23-2024 02:50 AM
hi I wanna measure 4-20ma current for flow rate using ni 9207 but when I measured, current graph was ploted constant function
like above graph
1. plz tell me what is wrong why current is over the 20mA and it did not fluctuate attached my block diagram
2. in the task making section there is external resistance what does that mean? I just connected ni 9207 and 4-20mA port
this connection is wrong? how can I know the resistance
this is circuit which I use
the main problem is current did not be measured
I wanna measure the values plz tell me the solution
if u need more information plz tell me
Solved! Go to Solution.
05-23-2024 03:27 AM
You connect the current signal to AI0+ and AI0- on the 9207. This is a voltage input. The current input is AI8 and you need to connect the current output between VSup (positive power supply voltage 12V or 24V depending on your sensor) and the AI8 input. And of course read the AI8 input in DAQmx.
05-23-2024 07:59 AM
holy I really appreciate it i don't know why I connected voltage hole
And could I ask u 1 more. what is external resistance in current task making section
flow meter device circuit resistance or should I connect another resistor?
05-23-2024 08:08 AM
The current measurement resistor is internal to the module and you do not have to do anything external in terms of a resistor. The documented input impedance for these channels is 85 Ohm.
05-23-2024 06:53 PM
thank u so much I solved my problems