cannot communicate using VISA

i'm a new user of labview 7.0 and trying to communicate with philips function generator PM5191. when i'm communicate using instrument io assistant, the front panel of the PM5191 is changing and i can get the output signal as i set in the string, but when i communicate using VISA, the front panel is changing but i cannot get the expected output from the PM5191. why?
0 件の賞賛
メッセージ1/9
3,298件の閲覧回数

Hi @yahoo.com,

are you sure your settings are correct? Start using shipped examples (Basic Serial Read and Write). Can you show your code or a picture of it?

Mike

メッセージ2/9
3,297件の閲覧回数
can u give me your email address, i'll email to you the file,maybe you can check it if there is some mistake
0 件の賞賛
メッセージ3/9
3,292件の閲覧回数

Hi @yahoo.com,

upload it here and everyone can help you. スマイリー ハッピー

Mike

メッセージ4/9
3,289件の閲覧回数
R u able to communicate using hyper terminal?
0 件の賞賛
メッセージ5/9
3,285件の閲覧回数
0 件の賞賛
メッセージ6/9
3,284件の閲覧回数
hi muks. i'm connected using gpib, so i don't know how to use hyperterminal.
0 件の賞賛
メッセージ7/9
3,275件の閲覧回数
this file using instrument io assistant.
0 件の賞賛
メッセージ8/9
3,274件の閲覧回数
First, you are not sending the same commands in the one with the VISA Write functions (look at the second VISA Write). Second, have you tried append a CR like you are doing in the assistant. Lastly, have you looked at the drivers for similar Phillips instruments? There's one for the PM5192 that might be similar to your 5191.
メッセージ9/9