Hi Every one

 

I'm new to labview. 

 

I use a high accuracy load cell and signal conditioner which provide precise excitation and output (10 V)

 

The problem that when I use the DAQ card (myDAQ) to measure the output voltage 

 

The voltage reaches zero for finite time as in the image however it is very stable at zero (at no load)

Also the same problem I found earlier when measuring a 9V DC battery

 

Any ideas?

 

0 Kudos
Message 1 of 9
(2,927 Views)

Looks like a wiring problem to me.

Chris
Certified LabVIEW Architect
Certified TestStand Architect
0 Kudos
Message 2 of 9
(2,916 Views)

All connections checked and good

 

aaaaaaaaaah very confused

0 Kudos
Message 3 of 9
(2,894 Views)

Please post a diagram showing exactly how everything is connected.  Also pst yoru VI with some typical data saved as default.

 

Lynn

0 Kudos
Message 4 of 9
(2,886 Views)

Moment please

 

my signal conditioner is programmed via serial port

 

when I connect it to pc via usb-232 converter in the same time of aqcuiring data then all readings are ok and no ripples or spikes to zero 

 

The connection diagram and VI block are attatched

 

 

 

 

Download All
0 Kudos
Message 5 of 9
(2,852 Views)
Solution
Accepted by topic author engcad

I suspect that you have a grounding problem. The signal conditioner is probably grounded through the USB/RS-232 connection when you have that connected. When it is not connected through the USB, it may not be grounded.

 

Try connecting the AGND of the MyDAQ to the 24 V DC- connection or to a ground connection on the signal conditioner.

 

Lynn

0 Kudos
Message 6 of 9
(2,836 Views)

thanks very much

0 Kudos
Message 7 of 9
(2,792 Views)

grounding the signal conditioner had solved the problem

0 Kudos
Message 8 of 9
(2,788 Views)

Thank you for letting us know.  When you report what solved the problem for you, it can help someone else with a similar problem.

 