Reading accelerometers using NI9230
Hello community. I am new to accelerometers and  am trying to measure vibration on rotating equipment. I have a B&K Vibro accelerometer (datasheet attached) . I connected it to an NI9230 and was able to get a signal out of it. Although I am not able to understand how to interpret what I am looking at. I would expect g values to show on y axis and show a constant ~1g when my accelerometer is vertical. Although what I see is values close to 0. It does respond to vibrations but I dont understand what I am looking at on the y axis. Do I need to scale the sensor readings? It does take 'g' as unit by default though.
Any help is appreciated! I have attached relevant snapshot.
Message 1 of 10
(4,825 Views)
Accepted by qusaimn
The acceleration due to gravity is "constant". If your accelerometer reads it, it measures a constant DC voltage. However, your accelerometer is AC coupled, which means it blocks DC voltages. That is why you are not seeing a constant of 1 g.
Message 2 of 10
(4,772 Views)
Thank you for the response! Your answer prompted me to search AC coupled accelerometers and I found this which makes perfect sense to me now.
Although I am still wondering what are the values I am seeing on the Y axis then? Is it just the amplitude of displacement of the vibrations then? How do I interpret that?
Message 3 of 10
(4,776 Views)
You are reading 1 sample at a 1 Hz rate. You are seeing noise and low frequency drift. Sample at 1000 Hz, read at least 100 samples. Tap the table; do you see a spike in the data? You may have to take at least 1000 samples to see the spike easily, you may miss it on a shorter  timescale. Bonus - take the FFT of the data, do you see any vibrations in the background?
Message 4 of 10
(4,757 Views)
Accepted by qusaimn
The accelerometer measures the vibration (accelerated motion) of an object. When the object is still, the output of the accelerometer is 0.
Generally, the vibration of a rotating part should below 0.1 g. If it exceeds 0.3 g, the vibration may be too large and the structure may be easily damaged.
Message 5 of 10
(4,734 Views)
That makes sense. I understand that. I was able to increase sampling rate and was able to see the spike.
However, what I am trying to understand is what is the units of the measurement on the y axis? Am I still measuring g forces(m/s2) or am I measuring the amplitude of whatever vibration I am detecting? Say I detect vibration spike in my data at 1Hz, what would that mean for the yaxis?
Message 6 of 10
(4,717 Views)
Accepted by qusaimn
You set the scale and conversion in DAQ Assistant, so the Y-axis should be units of g. This assumes the calibration is correct and the accelerometer was not damaged when dropped. The units of the Y-axis are always going to be g if you use the same settings in DAQ assistant. Mostly likely your accelerometer has a frequency dependent response curve, in the calibrated frequency span, the response is ~100mV/g. If 1 Hz is in this span, then it has the same response, if 1 Hz is outside of this span the response may be lower or higher, that is, 100 mV may not be 1 g. You need to look at the literature that comes with the accelerometer.
Message 7 of 10
(4,698 Views)
That makes perfect sense. Thank you for the explanation!
Message 8 of 10
(4,690 Views)
Accepted by qusaimn
All sensors I saw from Bruel&Kjaer (now HBK) came AT LEAST with a (production) calibration certificate (at least one frequency point).
The spec of your sensor say 100mV/g +-10%  and another +-10% in the frequency range 1.5Hz to 10kHz.
So without the individual cal-sheet value you can expect  10%  increasing to ~15% (20% worst case) to the lower end (High pass filter 0.5Hz) and higher end (resonance effects) of the frequency range.
With a calibration in the frequency range (usually 1/3 octave) you migth come down to 1-3% (depending on the cal-lab)  ... if needed.
Kodos for the well documented question!
If you use LabVIEW, have a look at the Sound&Vibration Tools.
Greetings from Germany
Henrik
LV since v3.1
Message 9 of 10
(3,219 Views)
I have posted another question on accelerometers - Accelerometer + LabView + Physics question - NI Community.
If you had any input for me I would greatly appreciate it. Thank you!
Message 10 of 10
(3,139 Views)
