Problem getting correct graph data

Solved!
Go to solution

OK, stop the program right after error occures and post this NI SPY log

0 Kudos
Message 11 of 20
(1,150 Views)

OK, the now the program stops when it gets an error.

 

In attachment: spy file, program

 

Thanks a lot for your time,

Thomas V.

Download All
0 Kudos
Message 12 of 20
(1,147 Views)

Is your VISA timeout wiat time less than your sweep time?

 

If possible, I always use some form of the OPC command (usually *OPC? because it's so easy to implement), and make the VISA timeout > longest sweep time.

Bill
CLD
(Mid-Level minion.)
My support system ensures that I don't look totally incompetent.
Proud to say that I've progressed beyond knowing just enough to be dangerous. I now know enough to know that I have no clue about anything at all.
Humble author of the CLAD Nugget.
0 Kudos
Message 13 of 20
(1,124 Views)

If the failure occurs, what do you see on the instrument screen? Can you see the correct trace data?

0 Kudos
Message 14 of 20
(1,108 Views)

Yes, when the query interrupted occurs, the traces on the device are projected correctly.

I looked at the raw data we get from the ZVL and 80% of the time the values are correct, the other 20% the values are false.

0 Kudos
Message 15 of 20
(1,102 Views)

It seems to be firmware related problem, I discussed it with development engineer. You're running FW 3.20 version, try to get the new 3.21 from R&S page:

 

http://www.rohde-schwarz.de/de/Produkte/messtechnik-testsysteme/netzwerk-analysatoren/ZVL13-|-Firmware-|-23-|-2584.html

0 Kudos
Message 16 of 20
(1,088 Views)

Ok thank you,

 

but the link you provided doesn't work, it says the page doesn't exist when I try to open it. Also tried searching for FW3.21 on the site but it gives 0 results.

 

 

0 Kudos
Message 17 of 20
(1,078 Views)

edit: the link works

0 Kudos
Message 18 of 20
(1,071 Views)

we've updated the drivers, but the problem is still here. Query interupted error and weird graphs.

0 Kudos
Message 19 of 20
(1,068 Views)
Solution
Accepted by Thomas_V

You will have to live with the reading the trace again on these ocasions. Fixing instrument FW will take some time. Use rszvl Clear Instrument Errors.vi from Utility group to clear this error from the instrument queue.