Incorrect pressure bridge sensor readings in NI9237 Module

I have connected Kyowa Pressure bridge sensors BEC-D-1MPA to a NI-9237.  These bridge sensors have an initial bridge unbalance 0.436 mv/V shown on the calibration sheet and a rated capacity of 0.315 mv/V at 1MPa.  The resistance of each sensor is 120.2 Ohm and these have 4-wires (2 input and 2 output). I have selected excitation of 2.5V and plugged in scaling of 0 mv/v@ 0 Pa and 0.315 mv/V @ 1E6 Pa.  The calibration constant is 0.318MPa / 1mv/v.  Using the example vi for pressure bridge sensors I select null offset, which zeros the initial reading. Unfortunately, as I apply increasing pressure I get readings with an increasing gain for each pressure. I am using only two ports of the NI-9237 so unlikely exceeding a power consumption limit of 150mW. I am unclear whether any jumpers are needed for the RS-/RS+ and SC connections as per the NI9237 instructions? Or this is handled by software null offset.

 

Not sure, how to resolve this. 

0 Kudos
Message 1 of 12
(346 Views)

I have connected Kyowa Pressure bridge sensors BEC-D-1MPA to a NI-9237.  These bridge sensors have an initial bridge unbalance 0.436 mv/V shown on the calibration sheet and a rated capacity of 0.315 mv/V at 1MPa.  The resistance of each sensor is 120.2 Ohm and these have 4-wires (2 input and 2 output). I have selected excitation of 2.5V and plugged in scaling of 0 mv/v@ 0 Pa and 0.315 mv/V @ 1E6 Pa.  The calibration constant is 0.318MPa / 1mv/v.  Using the example vi for pressure bridge sensors I select null offset, which zeros the initial reading. Unfortunately, as I apply increasing pressure I get readings with an increasing gain for each pressure. I am using only two ports of the NI-9237 so unlikely exceeding a power consumption limit of 150mW. I am unclear whether any jumpers are needed for the RS-/RS+ and SC connections? Or this is handled by software null offset.

 

Not sure, how to resolve this. 

0 Kudos
Message 2 of 12
(301 Views)

Please share the connection diagram.

 

You must connect all wires for best performance,

santo_13_1-1751603193076.png

 

Santhosh
Soliton Technologies

New to the forum? Please read community guidelines and how to ask smart questions

Only two ways to appreciate someone who spent their free time to reply/answer your question - give them Kudos or mark their reply as the answer/solution.

Finding it hard to source NI hardware? Try NI Trading Post
0 Kudos
Message 3 of 12
(312 Views)

Hello Santo13

The connection diagram and description are unclear as to what needs to be connected.  May I ask you to kindly clarify please?

Thank you.

0 Kudos
Message 4 of 12
(257 Views)

Hi Ox,

 


@OxSanj wrote:

The connection diagram and description are unclear as to what needs to be connected.


Which connection is "unclear" in the diagram?

Why do you ask for any "jumpers": there are no jumpers in the diagram…

Best regards,
GerdW


using LV2016/2019/2021 on Win10/11+cRIO, TestStand2016/2019
0 Kudos
Message 5 of 12
(244 Views)

Hello Gerd,

The full bridge sensors have only 4 wires - red (+excitation),  black (-excitation), Green (- Output signal) and White (+ output signal) which are connected to AI+, AI-, EX+ and EX-. The NI-9237 operating instructions and spec state "for a full bridge you must connect the AI+, AI-, EX+, EX and RS+ and -RS signals".  Kindly see Figure 3 of the NI-9237 (with RJ50). I have RS+ and RS- left unconnected, I am using an example vi with null offset and no shunt. But I end up with readings that require gain adjustment. Hence confusion prevails..

best wishes

Oxsanj

 

0 Kudos
Message 6 of 12
(236 Views)

Hi Ox,

 


@OxSanj wrote:
The NI-9237 operating instructions and spec state "for a full bridge you must connect the AI+, AI-, EX+, EX and RS+ and -RS signals". 

What happens when you connect RS+/RS- with EX+/EX-, respectively?

Best regards,
GerdW


using LV2016/2019/2021 on Win10/11+cRIO, TestStand2016/2019
0 Kudos
Message 7 of 12
(233 Views)

Hello Gerd

 

Thank you for your response. 

When I connect RS- with EX- and RS+ with EX+ the response of the pressure sensor is a flatline on the test panel.

OxSanj_0-1753351703381.png

 

Without the connections with RS+ and RS- the response of the sensor is as shown below 

 

OxSanj_1-1753351777098.png

 

The operating instructions state the following and provide the connection diagram. You must connect AI+ and AI-, EX+ and Ex-, and RS+ and RS- signals. However it fails to indicate what to connect with? Just poor wording...

OxSanj_4-1753353381223.png

 

OxSanj_2-1753352033751.png

The instructions indicate options to correct for resistance errors: remote sensing and shunt calibration. Per your suggestion I have followed Figure 4 and I get a flatline....

 

OxSanj_3-1753352623958.png

 

Thank you again for your guidance.

 

Oxsanj

 

0 Kudos
Message 8 of 12
(214 Views)

Hi Ox,

 


@OxSanj wrote:

The operating instructions state the following and provide the connection diagram. You must connect AI+ and AI-, EX+ and Ex-, and RS+ and RS- signals. However it fails to indicate what to connect with? Just poor wording...


All the connections are shown in Figure 3: how does that image "fail"?

 


@OxSanj wrote:

The instructions indicate options to correct for resistance errors: remote sensing and shunt calibration. Per your suggestion I have followed Figure 4 and I get a flatline....


Figure 4 is just a "boil down" of Figure 3 for those 6 connection lines…

 

Can you test your sensor with a simple power supply and a DMM?

Can you provide a datasheet of your sensor? I couldn't find one for the type number given in your first message.

 

(And please keep discussion in one place.)

Best regards,
GerdW


using LV2016/2019/2021 on Win10/11+cRIO, TestStand2016/2019
0 Kudos
Message 9 of 12
(207 Views)

Are you applying a varying pressure signal to consider a flatline reading from DAQ as bad?

 

If you're not applying any "varying" pressure signal, a flatline is a clean reading from DAQ.

Santhosh
Soliton Technologies

New to the forum? Please read community guidelines and how to ask smart questions

Only two ways to appreciate someone who spent their free time to reply/answer your question - give them Kudos or mark their reply as the answer/solution.

Finding it hard to source NI hardware? Try NI Trading Post
0 Kudos