The DAQ board I am using is a SC-2043-SG. I am trying to use the DAQ assistant to create a VI to measure strain from a strain gage, but every time I test the input, it gives me this error:
Error -200161 occurred at DAQ AssistantPossible
Reason(s):Device to which the sensor is attached does not have an
available internal excitation source.
Select another device with an
available internal excitation source or supply external
excitation.
Channel Name: Strain
Device: Dev1
This error was received when using an INTERNAL excitation source the board. I have the mx drivers installed, so I know that is not the problem. I the following
at this site:
SC-2042RTD and SC-2043SG:The SC-2042RTD and SC-2043SG can
both be used with NI-DAQmx. If you are using an RTD or Strain Gauge
Task or Channel you must choose external excitation and set it to the
values provided by these accessories. Also, the SC-2043SG provides a
gain of 10 to the strain gauge signal that must be taken into account.
So, I again tried to use the DAQ assistant, but changed the excitation source to external and a magnitude of 2.52V (that of the board). When this is done, there is no error received upon testing, but it does not give appropriate readings. Instead, it gives a straight line, and is unresponsive to changes in the applied load (the measurement was set to continuous).
I cannot figure out what I need to do in order to get this working. Any help would be greatly appreciated.