Hi,

 

I am trying to connect to a Keysight Infiniivision DSOX2004A Oscilloscope using LabView, without success. I am using a GPIB-USB cable from keysight, with the appropriate driver installed (82357B), although the usb-usb port isn't working either.

I use Labview 17, with NI-VISA 17.0, NI-488.2-17 and DCD-May17 installed in a Windows 10 - 64bit machine.

 

When I open Labiew and try to find instruments connected (Scan for instruments), the Instrument finder fails to detect the oscilloscope. The device doesn't show in NI-MAX neither...

 

The Keysight software can see the device, which is connected in GPIB0::7::INSTR, but Labview can not see it.

 

If I open any .vi from the Labview drivers of this scope (Agilent 2000 3000 X-Series), and run the .vi, I get an error code VISA error code: 1073807346. I expected to have the device listed in the drop-down menu of the VISA I/O Resource, but it isn't there, it only contains "Refresh".

 

I have tried to install and reinstall the software and drivers many times, but I still can't connect with the device, I don't know what else I can do.

 

Any help or ideas would we highly appreciated.

Many thanks,

 

ElTitoFranki

 

0 Kudos
Message 1 of 5
(6,677 Views)
Solution
Accepted by eltitofranki
I am using a GPIB-USB cable from keysight,

I had an Agilent USB-GPIB before and it needed the Tulip driver enabled to work

 

Open NI-MAX (Measurement and Automation Explorer)

 

On the menu bar click on Tools---->NI VISA----->Visa Options

 

Select Passports from the tree under General Settings

 

Make sure NiVisaTulip.dll (Ni-VISA Passport for Tulip) is checked

 

ppasse.PNG

========================
=== Engineer Ambiguously ===
========================
Message 2 of 5
(6,644 Views)

Thank you very much!

 

Now NI-MAX detect the device, although Labview doesn't show it in the drop-down menus in VISA I/O Resource..

Any ideas why?

I am completely new to LabView.

 

0 Kudos
Message 3 of 5
(6,627 Views)

I am now able to communicate via usb-usb cable, but only if I turn on the oscilloscope after I plug the cable.

 

0 Kudos
Message 4 of 5
(6,616 Views)

Have you solved the communication problem with your Keysight oscilloscope?? I have the same problem, but with Keysight or Agilent DSO1024A. I can't acquire the channels waveform in the Labview.

0 Kudos
Message 5 of 5
(6,266 Views)