instrument not found
Hi. I have an GPIB-USB-HS and i am trying to connect a TDS scope.Everything looks ok, but MAX can't find any instrument.
Message 1 of 10
(5,180 Views)
That's puzzling. I've never had a problem with TDS Scopes and GPIB before.
I assume you've already set an address through the Scope's menus. Other obvious questions - Are you using a known good GPIB cable?
Is this a single standalone instrument or are there other devices attached to the controller board?
Also I'm not sure about this, but I think the older TDS Scopes made you choose your preferred communication port (eg - Serial OR GPIB).
Message 2 of 10
(5,146 Views)
It's a NI GPIB. All i did was to instal the software from the cd. How do i set an address ?It's the first time i use an GPIB and probably there are lots of things i don't know.  The scope is a TDS 724.
Message 3 of 10
(5,128 Views)
Message 4 of 10
(5,110 Views)
Ssnakey--
Just out of curiosity, does the GPIB-USB-HS show up in MAX as well?  How about when you scan for instruments, do you get an error?  If so, what is it?
Tyler C
Message 5 of 10
(5,094 Views)
My GPIB driver version is : 2.5.2.1.
I've attached a printscreen with the way MAX see my hardware.Maybe there are some things i have to set for the scope.
Message 6 of 10
(5,052 Views)
Hi the screen you did sent is for the USB-HS GPIB interface and I'm wondering why it has address 1
Normally the gpib interface has address 0  and the instruments start to count from 1 up to 31.
I see indeed no instruments found, so that message is correct but what address does the scope have?
Message 7 of 10
(5,018 Views)
I've changed the GPIB primary address from 1 to 0 and still nothing.How can i find out what is the scope address?
Message 8 of 10
(5,000 Views)
Ssnakey,
Have you ensured that the device is set up for GPIB communications?  When you scan for instruments it will tell you the address of the device.  As Semper_discens said, some TDS scopes have to be either serial or GPIB.  If you have a way to change out cables (or a way to test this one with a known and working setup) that would be good to test as well.
Caleb W
Message 9 of 10
(4,979 Views)
Do you have the manual to see how to change the settings?  I only found a service manual, but that might help
http://energy.ece.illinois.edu/equipment/tds600c_service.pdf
Message 10 of 10
(4,956 Views)