06-19-2023 12:53 PM
The Initialize Vi is in this path
C:\Program Files\National Instruments\LabVIEW 2020\instr.lib\rssmb
06-19-2023 03:13 PM
While my path is
C:\Program Files (x86)\National Instruments\LabVIEW 2020\instr.lib\Rohde&Schwarz RF Signal Generator
So we clearly loaded different implementations.
Please try the LabVIEW 2015 installation. It works for higher versions as well.
You only should give it enough time to compile all vi's
06-20-2023 07:47 AM
Hellou,
At first, thank you for your attention. It is realy helping me in this task.
I get that you mannage to use the signal generator, but I have a vector signal analyser, thats why I'm using this other blocks.
06-20-2023 07:54 AM
If I try to use the Signal Generator block, I get this
06-20-2023 08:18 AM
Do you have SMBV-K131 option? If so, the current driver limits its use to SMW200A. New version coming at the end of Jun will fix that. In the meantime, do the following:
In the Configure Error Checking.vi switch the Option checking OFF. Then you're good to go.
Milo
06-20-2023 03:00 PM