Hi All,
I had NI HS-USB-GPIB Controller and keithley USB-488 Controller
Initially i was able to control instruments with NI HS-USB-GPIB Controller
Later i want to use keithley USB-USB Controller so i installed driver given by keithley USB-488 Controller
Now i want to revert back to NI GPIB controller
My problem i have uninstalled 488 GPIB driver given by keithley, installed NI USB-GPIB Controller drivers
I was able to see the controller and Instrument connected in the MAX, but i was not able to see it in labVIEW VISA control
could you suggest where problem could be
Thanks
‎04-09-2009 09:40 AM
Hello Appala,
I am unsure about what the problem could be. What versions of NI-488.2 and 488 GPIB were you using?
Also, you mentioned that you are able to communicate with the device (scan for instruments worked correctly), is this still correct?
Thanks,
Steven T.
‎04-09-2009 09:53 AM
‎04-14-2009 11:30 PM
Hi all,
Yes, i too suspect the same,
but i have unistalled "CEC-488", is there anything else to be uninstalled.
Thanks
‎04-15-2009 01:26 AM
OMG!
Like you can't define a VISA sesion to the device...
Keithley submits drivers to the NI web-site in (almost) real time
Primarilliy because the XXxx01 has a FW update "flash-port" abd is,(by mfg spec) "488.2 compliant " , goes to < gotos?> an unk state when it receives <"IDN?">
Don't *IDN? a kiethley instrument (It will hang up)
Don't *IFC while an unkown instrumet is hanging on GPIB:x (some controleres take this as "QU"- or "Quit Controlling"
READ the MFG users guides
DO NOT assume that a spec will tell you that a device isn't (really) 488.2
ANOTHER (488.2) NON-compliant device!
Anyone care to make a comprehensive list?
‎04-15-2009 07:32 AM
‎04-16-2009 07:28 AM
Hello Dennis,
I was able to see NI-VISA in MAX Software section