This error is coming from CMU driver. 0xBFFC908 error means that function group is not assigned to a secondary address. When you update k2g, did you update RSCMU base too? This can be possible source of troubles.
If you are sure that you have instelled the latest version of drivers, try contact Rohde&Schwarz. Send to RS version of you rscmu200_32.dll and rscmuk2g_32.dll, which are located in the [VXIpnp dir]\Winxx\bin.
The latest version should be
rscmu200_32.dll - 3.0.5.10 "Rev 3.05RC10"
rscmuk2g_32.dll - 3.0.5.10 "Rev 3.05RC10"
Try search rscmu*.dll on your system. These dll should be ONLY in [VXIpnp dir]\Winxx\bin.
E-mail to RS support is
CustomerSupport@rsd.rohde-schwarz.com
Good Luck,
Zdenek