Communicating with an Extech HD300 Anemometer

I'm fairly competent in labview and this will be my second time writing my own driver (although the first was more straight foward). The provided software from the manufacturer looks like it was written in labview so I asked if they could help me with some code and here's what was provided:

 

HD300 COMMUNICATION PROTOCOL 
Download protocol:
0	1	2	3	4	5	6	7	8	9	10
0xff	0xaa	0x55	k	data	X1	X2	X3	checksum	0xff
  (short) Checksum = k + data + X1 + X2 + X3
   Request to send the command: k = 0x63  data = 0x63   (X1,X2,X3 is arbitrary value)


Data table:k = ‘k’
	MAX/MIN(U)	AVG	HOLD(U)	IRT	UNIT	MAX/MIN(D)	HOLD(D)	AREA	POWER
Click	10	20	30		40	60	50	70	80
Double click	11				44	66			

IRT key click:  i = ‘i’
0	1	2	3	4	5	6	7	8	9	10
0xff	0xaa	0x55	i	10	X1	X2	X3	checksum	0xff

Upload protocol:
0	1	2	3	4	5	6	7	8	9	10	11	12	13	14	15	16	17
a	d	j	g	F1	F2	F3	F4	F5	Vel_mode	Ambient temperature	Air speed

18	19	20	21	22	23	24	25	26	27	28	29	30	31	32	33	34	35	36	37
IR temperature	DISP1	DISP2	DISP3	CFM AREA

38	39	40	41	       42    	43	44	45	46	47	48	49
CMM AREA	MULAVG COUNT	AREA INDEX	0xff	g	0xff	g	checksum

F1:
7	6	5	4	3	2	1	0
				F(U)	C(U)	MAX(U)	REC(U)

F2:
7	6	5	4	3	2	1	0
FLOW	VEL	TYPE K	IR TEMP	HOLD(D)	MIN(D)	MAX(D)	REC(D)

F3:
7	6	5	4	3	2	1	0
C(D)	AREA		PC-LINK	knots	X100	X10	AVG

F4:
7	6	5	4	3	2	1	0
km/h	Ft/min	m/s	CMM 	CFM 	ft2	F(D)	m2

F5:
7	6	5	4	3	2	1	0
			MPH	POWER	HOLD(U)	MIN(U)	


		else


vel_mode:  
0	1	2	3	4	5	6
m_s	ft/min	km/ h	mph	knots	CFM	CMM




 I'm not exactly sure what I'm supposed to do with this, it wasn't at all like what I was working with last time.

0 Kudos
Message 1 of 22
(5,595 Views)

I couldn't see if there's a way to edit posts, so I'll just add a bit more of relavent info. The previous time I wrote a driver I used the Instrument I/O Assistant and used serial commands for the device as refrence. Not even sure what the above table is.

0 Kudos
Message 2 of 22
(5,589 Views)

It's the list of serial commands. It appears you have to send/receive data as a series of hex bytes (prefix, data, checksum, suffix). With the Byte Array to String, you can create the command and pass it to the VISA Write function. There have been numerous posts on using hex data and checksum calculations.

 

I will never use the Instrument I/O Assistant as it does not create a 'proper' instrument driver. I have no idea if it can be used in this situation.

Message 3 of 22
(5,581 Views)

Well, I've since rewritten my previous driver properly using VISA to try and understand better, but this protocol is substancially more complicated than my previous one.

 

I'm not sure if I should be writing the upload protocol or the download protocol.

 

It looks like the IRT key click (infrared thermometer) command is given as an example, except byte 3 which is still defined as 'i' which I don't understand.

 

And X1,X2,X3 can be any value for the purposes of the checksum?

 

I've added a picture of the VI as I'm working on it.

0 Kudos
Message 4 of 22
(5,568 Views)

You may have to write both upload and download. I have not studied it in detail, but if you need to send and receive, then you will need both. And, if the send commands will change based on some user option, then you will need to dynamically create the checksum.

 

For your testing purposes, I think it is easier to just start with a U8 byte array instead of all of those conversion functions (which are incorrect as I8) and the build array but in the long run, you will probably need both.

Message 5 of 22
(5,557 Views)

If the manufacturer's code is written in LabVIEW then perhaps they will share the actual LabVIEW source code that they use for their application? I'd try pushing on them a little bit more as you're essentially just going to be redeveloping what they've already done if their code truly is developed in LabVIEW.

 

Lars L

 

Message 6 of 22
(5,540 Views)

If only...I asked if there was anything else they could provide, source code, examples, documentation. That's all I got. They assume you either know how to use the above or just use their software.

 

The provided software gives an interactive diagram of the device where you can press any button and it works on the diagram and on the device. So, I'm trying to simply send the command to use the built in Infrared Thermometer (it beeps) and not even recieve data. But I don't think I'm coming up with the proper command, haven't managed any progress yet.

0 Kudos
Message 7 of 22
(5,538 Views)

Have you tried using a serial port sniffer such as portmon from Microsoft. If their code was truly written in LabVIEW, you can used NI-Spy/NI-Trace to capture the commands.

Message 8 of 22
(5,535 Views)

Thanks for the advice for sniffing with Portmon, it took some resource hacking to get it to work properly (apparently a known issue, but the dev seems to have abandoned it).

I used it in conjunction with the included software (built in VC++ considering the dll I found in there, not Labview 😞 ) and got way more information than I bargained for. I stripped out all the commands that I think were simply from the serial to usb.

 

Additionally I stripped out what I believe to be the ACK writes to the device (length 1: 1) so as to better compare each read.

 

As the log name suggests, the only thing I wanted to get out of this was to read the ambient temperature as it appeared on the device and software. According to the protocol above, this should appear between the 10-13 bits, but there is an additional bit AA at the beginning so I believe it's actually 11-14th bits. Taking the first reading, C2 F9 B2 41, this number is nowhere near 72.4F in decimal, so I'm not sure what's going on. The protocol has a way of knowing if it is set to F or C (the F1 bit at 4+1 offset, which reads as 08, not on the protocol), so I don't see why it would be in kelvin or such and the conversion would be software side.

 

The only things I've managed to confirm from the protocol and the software is that Vel_mode is set to ft/min = 1 as in bit 9+1offset. All other fields seem to be giving values that aren't listed on the documentation.

0 Kudos
Message 9 of 22
(5,517 Views)

At this point, it couldn't hurt to try and ask the manufacturer for help again. I know it's been difficult, but you've got more information on the port communication and a more specific problem description, so hopefully they'll be a bit more generous.

KyleP
Applications Engineer
National Instruments
Message 10 of 22
(5,500 Views)