| Instrument | Recommended Model | Requirements | 
|---|---|---|
| Calibrator | Fluke 5700A | If that instrument is unavailable, use a high-precision voltage source that is at least 50 ppm accurate for 12- and 13-bit boards and 10 ppm for 16-bit boards. | 
| DMM | NI 4070, Agilent 3458A | If these instruments are unavailable, use a multi-ranging 5.5-digit DMM with an accuracy of 40 ppm | 
| Counter | Agilent 53131A | If this instrument is unavailable, use a counter with an accuracy of 0.01%. | 
| Chassis | NI PXI-1042, NI PXI-1042Q | Use with PXI modules. | 
| DAQ Accessory | NI SCB-68 | Shielded I/O connector block with 68 screw terminals for easy signal connection to 68- or 100-pin DAQ devices. | 
| NI CB-68LP, NI CB-68LPR, NI TBX-68 | Low-cost termination accessories with 68 screw terminals for easy connection of field I/O signals to 68-pin DAQ devices. | |
| Shielded DAQ Cable | NI SH68-68-EP | Use with NI 671x/672x/673x modules with a 68-pin SCSI II connector. | 
| NI SHC68-68-EP | Use with NI 6715. | |
| NI SH68-C68-S | Use with NI 6722/6723. | 