Multifunction DAQ
IEPE Accelerometer Connectivity check on NI 9234
I am using a 4 channel NI 9234 Analog IEPE module with a cDAQ NI 9184 chassis. I have an IEPE accelerometer and an IEPE Microphone on the first two channels of the NI 9234 module. I am using this setup for a test DC motor test system. I would like to check if the accel and the mic are connected before every test programmatically. Is there any NIDAQmx function to acheive this? Can I check if an IEPE sensor is connected to an Analog channel before starting the data acquisition task??
(3,691 Views)
the common solution is to read the bieas voltage in DC mode, but that doesn't seems to be possiblewith this unit ??
Ways around:
The best way: If the sensors support TEDS , use it !  You can even identify the sensor 🙂
If TEDS doesn't work: (and open sensor and short detection is not supported)you can try this:
turn off IEPE current
switch to dc for some ms.  (Discharges the AC coupling cap)
switch back to AC and turn on IEPE current and immediatly start reading values. If a IEPE sensor is connected you should see a ~10V peak drifting to zero.
Greetings from Germany
Henrik
LV since v3.1
(3,668 Views)
Yes, a NI application engineer confirmed that reading the biased voltage is not supported on this module. I will try TEDS first. Thanks for the reply!!
(3,596 Views)
