Error -200478 - Specified operation cannot be performed when there are no channels in the task
I am writing a task in NiDaqmx to read 8 thermocouples from 2 NI 9222 and 4 thermocouples from a NI 9223 modules. Under CJC Source I have entered channel and specified a K-type thermocouple connected to a NI 9219. All modules are located in the same cDAQ9178 chassis.
I get the above error.
I add the CJC channel and it complains that I have duplicate readings of the same channel in the one task.
The environment where this will be installed will vary greatly thus setting the CJC to Constant is not a viable option.
Any help would be appreciated.
(4,311 Views)
Here are some screenshots that may aid in the understanding of what I have configured
(4,296 Views)
Greetings,
Is there any specific reason why you are using NI 9222 and NI 9223 modules to read your thermocouples?
If you have the option to use a NI 9213 (http://sine.ni.com/nips/cds/view/p/lang/sl/nid/208788) module instead, it would be much easier (and cheaper).
The NI 9213 module can read 16 thermocouples at the same time and also has a built in CJC source, thus you would only use one module instead of four.
That being said I tried your configuration with simulated hardware in MAX and the task did run, but I didn't get any values out of it. I don't know if this is because of simulated hardware or there is another reason.
Unfortunately I don't have the 9222 and 9223 modules to test it with real hardware.
Attached is the VI I used (LV2010).
Best regards
Matej
/* A smart device is only as smart (stupid) as its programmer. */
(4,295 Views)
I think the problem is that you put the NI9219 channel on the bottom of the list.
Try putting it on top so you read that channel first. I did the same in my VI.
Matej
(4,288 Views)
Accepted by Le_Gross_Bill
Simple reason - we already own the NI 9222s and the NI 9223. I'll check the NI9213 prices and see if the group has the budget.
I looked at your program and compared it to mine. Tried to select the CJC channel as Temperature_12 but all local channels were greyed out. BUT I moved Temperature_12 up the order to the top as you had and Voila! it's available to be selected. I set all the 9222s and 9223 channels to to point to Temperature_12 and it saved without error.
So I'd say that putting the CJC channel as the first channel to read fixes the problem - this makes sense as the subsequent channels need a reading to use as their CJC. Where that logic breaks down is when you consider that the 9222 and 9223 channels are simultaneous channels but that is a topic for another post!
For now, thank you very much for your assistance.
Dean
(4,286 Views)
Glad I could help.
Don't forget to mark your post as solved.
Best regards
Matej
(4,282 Views)
