03-04-2008 01:34 PM
03-06-2008 01:13 AM
Hi Jay
I am assuming that you are using the NI-9237. This strain module has only a specific set of sampling rates. If you specify a rate that is not supported it will corers the rate to the next higher sampling rate. Also since you are using the cDAQ chassis, there is only sample clock so it must choose the highest sampling rate. So it is not possible to sample at a lower rate. Please see the following link.
If you are limited to Signal Express, this task is tricky. If you select a N-Sample acquisition, set the samples to read to greater than two, and set the post acquisition delay under the execution control tab to 1000, you acquisition will acquire the two samples at the very fast rate, but it will wait 1000mS between each acquisition. This will slow the overall sample rate.
Chris_K