 Yltsä
			Yltsä
			11-30-2021 02:25 AM
Hi, is it possible to use NI SCC-TC01 and SCC-TC02 modules in the newest DAQmx sw?
We have made a lot of trials but without any success. The other SCC modules are OK.
At the moment I have NI USB-6255 and SCC-68 adapter board. Please can you give any tips how to use SCC-TC0X thermocouple modules?
BR,
Ilkka
 VA.KI
			VA.KI
			12-06-2021 06:51 AM - edited 12-06-2021 06:52 AM
According to NI-DAQmx 21.3 Readme the NI-TC01 is supported but I think it is related to USB version of it.
Which version of DAQmx you are using?
12-06-2021 08:29 AM
I use DAQmx 21.3. I have also the USB-TC01. That works OK, but we need to make continuous buffered data acquisition by using the analog inputs.
The old SCC modules would be very practical for us.
BR,
Ilkka
 VA.KI
			VA.KI
			12-14-2021 06:09 AM
I've checked the documents.
I wonder what is the issue in our case?
The problem is a physical connection or you are getting an error on LabVIEW?
12-14-2021 06:26 AM
Hi, There are just some small millivolts in the analog input. Cannot get right scaled temperature values out from the SCC module. BR, Ilkka
 VA.KI
			VA.KI
			12-14-2021 07:01 AM
Can we be sure that it is not a hardware issue?
12-14-2021 10:43 AM
yes.. it can be HW issue too. We have tested several SCC modules and data acquisition cards. But the thermo element measuremets are not OK by using the SCC-TC0X modules.
 VA.KI
			VA.KI
			12-15-2021 12:40 AM
So, I'm afraid that the issue is most probably in the hardware.
I would recommend having it sent for repair but as the hardware is old I'm not sure that the repair service will work for it now.
It is better to renew the hardware as a solution.
12-15-2021 12:45 AM
We have tested several SCC-TC01 and SCC-TC02 modules and three different mother boards for those.
The other SCC modules are OK (strain gage, feed throught), the problem is just SCC-TC0X.. we have those 5 pieces..