Signal Conditioning
Custom Scaling Issue
Hardware: I have a SCXI - 1000 being controlled by a SCXI 1600 module. I am reading a voltage into a SCXI-1100 with a SCXI-1303 as a terminal block.
Software: SignalExpress 2.5
OS: Windows XP
Signal Input range is set at -10 to 10 Volt range. I tried to setup a linear custom scale =>
Temp = 233.4(Voltage) - 18
With this scale the temperature maxes out at 28.678, but if I take the scale off the voltage is not capped and keeps moving. Can anyone tell me what is going on?
(6,031 Views)
Is this response you get in Signal Express, MAX, or both?
(6,026 Views)
Both
(6,023 Views)
Hi Jack
Let see if your temperature maxes out at 28.678 that will correspond to a voltage of 0.199 and can infer from your post that without the scale you are able to read more than 0.199 volts.
I have two solutions for you: first you can change the “Signal Input Ranges” to a range including the result in temperature that you are expecting, see how the “scaled units” changes from volts to “custom”. Second you can create a formula to scale the volts to temperature and avoid changing the “Signal Input Range” to the new custom scale.
I hope it helps
Message Edited by Jaime F on 11-07-2007 05:54 PM
Jaime Hoffiz
Product Expert
(6,018 Views)
The issue was the limits on the custom scale. Thank you Jane. 😄
(6,008 Views)
