Hi all,

I have a Dytran 3200B IEPE accelerometer installed on a plant that must be connected to a NI-9234 module. The distance between the acceletometer and the module is about 50 meters.

 

My doubts are if the accelerometer will work with the 9234 module. The NI module has an IEPE excitation current of 2mA, the Dytran accept a power supply from 2 to 20mA and the bias voltage is from 7.5 to 9.5 VDC. The 2mA will be enough considering also the distance?

 

I attach also the datasheet of the accelerometer.

 

Thanks

Message 1 of 3
(113 Views)
Solution
Accepted by topic author Penzium

Hello My Friend

Yes its compatible up to 300 Meter to connect IEPE and to make sure review this link so you get more data

 Using Long Cables with IEPE Sensors - NI

 

for any help contact me 

Ahmedshafik445599@gmail.com

CLA
Message 2 of 3
(101 Views)

Hello, thanks for the reply and the link. Very useful.

 