01-13-2014 09:20 AM
Hello everyone,
I deleted by mistake some files from my PXI. Now I cannot boot my PXI properly. I always get a error message as " Error reading Ethernet Configuration File c:\ni-rt\...". I wish I could format my PXI from the MAX but as long as the ethernet doesn't work I cannot communicate it on Windows. The only thing I can do is to start the Bios Setup pressing F10 when the PXI starts.
What is the fastest and less traumatic way to restore my PXI?
Thanks,
Juan Carlos
01-13-2014 10:35 AM
Enter the BIOS, and set the LabVIEW Boot Configuration to "LabVIEW RT Safe Mode" - that will prevent the system from booting the software on the controller, and will instead boot into Safe Mode. Then you can reinstall software on the RT controller via MAX, and after reinstalling software you can change the LabVIEW Boot configuration back to "LabVIEW RT".
-Danny
01-13-2014 10:45 AM
Hello Danny,
Thanks for your answer. I tried what you said, but even doing that I cannot see my PXI from the MAX. I think it is because I deleted one file requiered to perform the ethernet connection. The best option I see is to format my PXIe-8108 directly from the BIOS, but I don't know yet how to do this.
Thanks again,
Juan Carlos
01-13-2014 10:50 AM
You can't format the PXI controller from the BIOS.
After setting the PXI controller into Safe Mode, and rebooting the controller, did your PXI controller get an IP address? Was the IP address a valid DHCP IP address, or was it in the form 169.254.xxx.yyy?
Is your system with MAX on the subnet that the controller is on? MAX can only see devices within the same subnet as the PC MAX is running on, unless you manually create an entry in MAX for the remote target on an external subnet.
-Danny
01-13-2014 11:09 AM
Hello,
I got the IP Address 169.254.115.81/16. If I press Refresh on MAX my device still doesn't appear. I tried to create a new device using the options: FieldPoint Ethernet, Remote Device and Remote VISA System, but in all cases the MAX was unable to find the PXI.
I was thinking about formating the PXI by using a USB memory.
Regards,
Juan Carlos
01-13-2014 11:09 AM
Hello,
I got the IP Address 169.254.115.81/16. If I press Refresh on MAX my device still doesn't appear. I tried to create a new device using the options: FieldPoint Ethernet, Remote Device and Remote VISA System, but in all cases the MAX was unable to find the PXI.
I was thinking about formating the PXI by using a USB memory.
Regards,
Juan Carlos
01-13-2014 11:22 AM
Juan Carlos,
169.254.115.81 is a bogus IP address; it's in the "Link Local" range, which basically means the network subsystem "gave up" trying to communicate with a DHCP server and instead issued a link-local IP address in case the network topology you have is using Link Local addressing (the vast majority do not).
Does this PXI controller have multiple network interfaces? You want to make sure you plug the ethernet into Port 1 on the controller if so, because the networking on a PXI controller is NOT redundant; you cannot just "use any network interface" - Safe Mode is configured to use Port 1 exclusively.
-Danny
01-14-2014 04:39 AM
Hello,
I checked what you said, but my PXI only has one Ethernet interface. I don't know what else to try.
Thanks,
Juan Carlos