Multifunction DAQ
what is the gain of usb-6353 DAQ when input range is between -100 to 100 mV?
simple question I know, but would really appreciate if someone can tell me the gain of usb-6353 DAQ when input range is between -100 to 100 mV.
Well, why not gains of the complete input range then...
–10 V to 10 V
–5 V to 5 V
–2 V to 2 V
–1 V to 1 V
–500 mV to 500 mV
–200 mV to 200 mV
–100 mV to 100 mV
Many thanks in advance!
Sincerely,
Hemm
(2,716 Views)
Hemm,
Gain is a relative term.  The X Series Users Manual (page 4-1) indicates that these devices use the NI-PGIA (programmable gain instrumentation amplifier) between the input multiplexer and the A/D converter.  It also indicates that the PGIA can amplify or attenuate.  Since the input range of the A/D converter is not specified, there is no way to tell which input ranges use amplification and which use attenuation.
Compared to the +/-10 V range the +/-100 mV range is amplified by a factor of 100.
Lynn
(2,697 Views)
Hello Hemm,
Please let me know if you have any further questions about this.
I think I also received an e-mail from you in our inbox.
A correction to my reply in that e-mail: The [-0.1V, 0.1V] range is supported. (I must have been looking at another datasheet that day)
Kind Regards,
Thierry C - CLA, CTA - Senior R&D Engineer (Former Support Engineer) - National Instruments
If someone helped you, let them know. Mark as solved and/or give a kudo. 😉
(2,676 Views)
