We have  a PCI-6036E DAQ board, use driver Traditional NI-DAQ (Legacy) C, Version 7.xx. Currently the input range is set to -+5V use function

SCAN_Setup (deviceNumber, numChans, chanVector, gainVector).

The gain vector is set as  

short piGainVect[8] = {1,1,1,1,1,1,1,1};

 

Now we want to set input range to -+10V. Is it posible and how to do it?

 

Thanks!

 

Jiachun

0 Kudos
Message 1 of 2
(3,095 Views)

Hello Jiachun,

 

Thanks for using NI forums.  The PCI-6036E has four software selectable input ranges, one of them being +/-10V.  According to the E Series User Manual, a gain of 0.5 will give you an input range of +/-10V.  All of the gain settings require an integer value, but according to the DAQ Device Gain Settings table (in the Traditional NI-DAQ (Legacy) C Function Reference Help) you need to use a value of -1 in order to receive a gain setting of 0.5.  The Traditional DAQ help file can be found by going to Start>>Programs>>National Instruments>>NI-DAQ>>Traditional NI-DAQ Function Reference Help.  This should work for you but please let me know if you run into any problems.

 