Multifunction DAQ
'output impedence' and 'voltage stability' on 673x series
  I'm trying to find some spec info and didn't see it on the website.  I was hoping to get the "output impedence" and the "voltage stability" for the PCI 673X series and the PCI 6703.  I have a customer and they are driving electron lenses and need very high stability.  They need at least 8 16 bit channels.  Hope you can help.
Thanks!
(2,722 Views)
Hello DAQ Info,
Thank you for posting to the discussion forum. I see that you are looking at the specs for PCI-673x and PCI-6703.
For the specs for the PCI-6703 reference the following user manual for the part.
NI 6703/6704 User Manual
The output impedance of the voltage output of is 0.1 ohms max. The stability is listed in A-2 of the user manual.
Looking at the specifications the output impedance is also 0.1 ohm.
Please let me know if something is unclear or if something is missing.
  Sandra T.
Applications Engineer | National Instruments
(2,698 Views)