hello,
If the USB 6008 cannot support hardware clock generation, then is there any means to do the real time data acquisition?
I wish to read in real time EMG sensor signal and then display it in MATLAB via the USB 6008 hardware, I manage to display real time input signal in MATLAB however facing problem displaying the analog output signal.
08-21-2007 09:30 PM - edited 08-21-2007 09:30 PM
hi Rob,
Thanks for your advise.
I had tried out some examples, the USB-6008 can do START/STOP/GETDATA operations, but it is having problems with the PUTDATA/PUTSAMPLE operations. Does this mean that i have to use other DAQ hardware in order to carry out those operations?
Regards,
yeelin
Message Edited by cat_e0 on 08-21-2007 09:33 PM
08-22-2007 12:57 PM
Hello Yeelin,
The USB-6008 is a low cost solution and only supports software timed generation. (USB-6008 Manual pg.16) Abbinav is correct; the maximum sampling rate is 150S/s. You currently have the sampling rate set to 10,000S/s. Are you able to get your USB-6008 to work correctly in Measurement & Automation Explorer (MAX)?
Any of our M-Series boards will be able to provide hardware timed generation. Here is a great forum post that describes using DAQmx and M-Series DAQ with MATLAB®.